Title: Hantek 6025BE calibrate [Print this page]

Author: LKIJ    Time: 2016-11-22 23:27
Title: Hantek 6025BE calibrate
Hi. I buy Hantek 6025BE osciloscope. And i have three question:
1) How need calibrate CH1 and CH2 ZERO (Utility -> Calibration)? Need short BNC socket in osciloscope panel or need connect probe to BNC socket and short ends of probe? Because probe center wire has ~250 Om/8,9 MOm (x1/x10) resistance. Calibrate must be done in x1 or x10 mode?;
2) How need calibrate trigger ZERO, as well as CHx in first question?;
3) Calibrating time? CH1 and CH2 was calibrated 2 hours, TRIG i will test this night, because i see need many hours.



Author: amy    Time: 2016-11-23 11:16
Sir, when the channel zero level offset or trigger offset, you can calibrate it.
When calibrating, disconnect any probes or cables from the input connectors.
This process only takes a few minutes.

Author: LKIJ    Time: 2016-11-23 16:53
Who influences long calibrating time? TRIG calibrating after 15 hours was not ending. Now i test on other PC.

Author: LKIJ    Time: 2016-11-24 17:23
Dear sir, TRIG calibrating never ending. Then i try measure (CH1) reference 1 kHz signal, from osciloscope panel whit trigger source CH1, then whit external trigger source, when in EXT input connected 1,3 V battery, but i not see signal line in application. Autoset function also not show signal line.
Is it damage and i must return product back to seller www.tme.eu ?

Author: amy    Time: 2016-11-28 14:28
LKIJ replied at 2016-11-24 17:23
Dear sir, TRIG calibrating never ending. Then i try measure (CH1) reference 1 kHz signal, from oscil ...

Sir, could you solder?
If you can, please replace the component D406(A DIODE, MBR0520,1R1C).
[attach]1266[/attach]


Author: LKIJ    Time: 2016-12-1 23:18
Thank you for answer. I already return osciloscope to seller.