NULL Reading

The DMM allows a separate null setting to be saved for the temperature function. When making
null measurements, each reading is the difference between a stored null value and the input signal.
One application of NULL is to increase accuracy of two-wire resistance measurements by first nulling
the closed–circuit test lead resistance.

Autozero On/Off

Enabling the autozero feature (ON) provides greater accuracy; however, the additional measurement
(of zero) reduces the reading speed.


58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
192#
 Author| Post time 2024-5-10 08:26:15 | Show all posts
Making High–Speed AC Measurements

The multimeter's AC voltage and AC current functions implement three low–frequency filters.
These filters allow you to trade–off minimum measured frequency for faster reading speed.
The FAST filter settles in 0.025 seconds, and is useful for frequencies above 200 Hz. The
MEDIUM filter settles in 0.625 seconds for voltage and 0.25 seconds for current, and is useful
for measurements above 20 Hz. The SLOW filter settles in 2.5 seconds for voltage and 1.66
seconds for current, and is useful for frequencies above 3 Hz.


58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
193#
 Author| Post time 2024-5-13 08:15:29 | Show all posts
Edited by WisdomAugust at 2024-5-14 08:37


With a few precautions, you can perform AC measurements at speeds up to 500 readings
per second.

Use manual ranging to eliminate auto-ranging delays. By setting the trigger delay to 0, the
FAST, MEDIUM, and SLOW filters allow up to 500, 150, and 50 readings per second, but with
reduced accuracy because the filter may not fully settle. If the sample–to–sample levels are
similar, little settling time is required for each new reading. Under this specialized condition,
the MEDIUM filter provides reduced accuracy results at 20 readings per second, and the FAST
filter provides reduced accuracy results at 200 readings per second.



58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
194#
 Author| Post time 2024-5-14 08:41:30 | Show all posts


In applications where sample–to–sample levels vary widely, but the DC offset level does not change,
the MEDIUM filter settles at 2 to 4 readings per second (depending upon the lowest frequency
component in the waveform) as shown in the following table:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
195#
 Author| Post time 7 day(s) ago | Show all posts
Edited by WisdomAugust at 2024-5-16 08:10

For AC voltage, additional settling time may be required when the DC level varies from sample to
sample. The default sample delays allow for a DC level change of 3% of range for all filters. If the
DC level change exceeds these levels, additional settling time is required. The multimeter's DC
blocking circuitry has a settling time constant of 0.2 seconds. This settling time only affects
measurement accuracy when DC offset levels vary from sample to sample. If maximum
measurement speed is desired, you may want to add an external DC blocking circuit for circuitry
with significant DC voltages present. This circuit can be as simple as a resistor and a capacitor.


For AC current, additional settling time is not required when the DC level varies sample to sample.



58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
196#
 Author| Post time 6 day(s) ago | Show all posts
Making High–Speed DC and Resistance Measurements

To make the fastest (but least accurate) DC or resistance measurements:

l Set the integration time (NPLC or aperture) to minimum
l Select a fixed range (autorange off)
l Disable autozero
l Disable offset compensation (resistance measurements)

Refer to the particular measurement type in Measurements for more information on the above
functions.


58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
197#
 Author| Post time 5 day(s) ago | Show all posts
Capacitance

The multimeter makes capacitance measurements by applying a known current to charge the
capacitance and then a resistance to discharge as shown below:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
198#
 Author| Post time The day before yesterday 11:41 | Show all posts
Edited by WisdomAugust at 2024-5-21 08:10

An illustration of the response curve while charging is shown below:




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
199#
 Author| Post time Yesterday 08:12 | Show all posts
Capacitance is calculated by measuring the change in voltage (DV) that occurs over a “short aperture”
time, (Dt). This measurement is repeated at two different times during the exponential rise that occurs.
An algorithm takes the data from these four points, and by linearizing that exponential rise over these
“short apertures”, accurately calculates the capacitance value.
The measurement cycle consists of two parts: a charge phase (shown in the graph) and a discharge phase.
The time–constant during the discharge phase is longer, due to a 100 kΩ protective resistor in the measurement
path. This time–constant plays an important role in the resultant reading rate (measurement time). The
incremental times (or “sample times”) as well as the width of the “short apertures”, vary by range, in order
to minimize noise and increase reading accuracy.

For the best accuracy, take a zero null measurement with open probes, to null out the test lead capacitance,
before connecting the probes across the capacitor to be measured (see Capacitance Measurements for details).


58

Threads

1279

Posts

1359

Credits

金牌会员

Rank: 6Rank: 6

Credits
1359
200#
 Author| Post time 9 hour(s) ago | Show all posts
Capacitance Measurement Considerations

Capacitors that have a high dissipation factor or other non-ideal characteristics will affect capacitance
measurements. Capacitors with high dissipation factors may exhibit a variance between the measured
value using the multimeter versus the single frequency method of some other LCR meters. The single
frequency method will also see more variation at different frequencies. For example, some inexpensive
capacitance substitution boxes, when measured with the multimeter, are almost 5% different compared
to the same capacitance measured with the single frequency method of an LCR meter. The LCR meter
will also show different values at different frequencies.

Capacitors with long time constants (dielectric absorption) will result in slow measurement settling time,
and will take a number of seconds to stabilize. You may see this when first connecting a capacitor or
when the delay time to make a measurement is varied. A high quality film capacitor typically shows the
least of this and an electrolytic capacitor the most, with ceramic capacitors typically in between.


You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-22 18:19 GMT+8 , Processed in 0.171692 second(s), 21 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list