Electronic Engineer Discuss

Author: WisdomAugust
Print Prev. thread Next thread

Digital Multimeter HDM3055 Series Manual

[Copy link]

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
51#
 Author| Post time 2023-6-13 08:03:46 | Show all posts
Step 5: Press AC Filter and choose the filter for the measurement. The instrument uses three
different AC filters that enable you either to optimize low frequency accuracy or to achieve faster
AC settling times following a change in input signal amplitude.
The three filters are 3 Hz, 20 Hz, and 200 Hz, and you should generally select the highest frequency
filter whose frequency is less than that of the signal you are measuring, because higher frequency
filters result in faster measurements. For example, when measuring a signal between 20 and 200 Hz,
use the 20 Hz filter.
If measurement speed is not an issue, choosing a lower frequency filter may result in quieter
measurements, depending on the signal that you are measuring.


NOTE For accurate displayed statistics of AC measurements in Front Panel mode, the default manual
trigger delay ([Acquire] > Delay Man) must be used.

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
52#
 Author| Post time 2023-6-14 08:30:52 | Show all posts
Resistance
This section describes how to configure 2-wire and 4-wire resistance measurements from the front panel.
Step 1: Configure the test leads as shown.
2-wire Resistance:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
53#
 Author| Post time 2023-6-15 08:20:40 | Show all posts
[size=11.0000pt]4-wire Resistance:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
54#
 Author| Post time 2023-6-16 09:01:56 | Show all posts
Step 2: Press [Ω2W] or [Ω4W] on the front panel. The following menu appears. (The Ω4W menu
does not include Auto Zero.)


Step 3: The Aperture NPLC softkey is selected by default. Use the up/down arrow keys to specify
integration time in power-line cycles (PLCs) to use for the measurement. 1, 10, and 100 PLC provide
normal mode (line frequency noise) rejection. Selecting 100 PLC provides the best noise rejection
and resolution, but the slowest measurements.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
55#
 Author| Post time 2023-6-17 08:18:20 | Show all posts
Step 4: Press Range to select a range for the measurement. (Auto (autorange) automatically
selects the range for the measurement based on the input. Autoranging is convenient, but it results
in slower measurements than using a manual range. Autoranging goes up a range at 120% of the
present range, and down a range below 10% of the present range.

Press More to switch between the two pages of settings.

Notice the amount of test current sourced is shown for each range. After selecting a range, the main
resistance menu is displayed.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
56#
 Author| Post time 2023-6-19 09:18:01 | Show all posts
Step 5: Auto Zero: Autozero provides the most accurate measurements, but requires additional
time to perform the zero measurement.With autozero enabled (On), the DMM internally measures
the offset following each measurement. It then subtracts that measurement from the preceding
reading. This prevents offset voltages present on the DMM’s input circuitry from affecting
measurement accuracy. With autozero disabled (Off), the DMM measures the offset once
and subtracts the offset from all subsequent meaurements. The DMM takes a new offset
measurement each time you change the function, range, or integration time. ( There is no
autozero setting for 4-wire measurements.)


58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
57#
 Author| Post time 2023-6-20 09:08:48 | Show all posts
Negative Resistance Measurements
Under certain conditions, the instrument may report negative resistance measurements. These may
occur in 2-wire and 4-wire resistance measurements or continuity measurements.
Conditions that may cause negative ohms values include:
l Changes in Front/Rear switch contact resistance
l Reversed Sense Hi and Lo leads
l Circuits with external bias or thermal voltages at circuit connections
l Changes in measurement connection after a NULL operation
Under the same conditions, the DMM returns the measurement’s absolute value so as to prevent
confusion associated with negative readings. The Hantek HDM Series DMMs will return negative
values. This allows the most accurate results after a NULL operation.


58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
58#
 Author| Post time 2023-6-21 10:01:11 | Show all posts
Temperature
This section describes how to configure 2-wire and 4-wire temperature measurements from the front panel.
Temperature measurements require a temperature transducer probe. The supported probes are 2-wire and
4-wire RTDs, 2-wire and 4-wire thermistors.
Step 1: Configure the test leads as shown.
2-wire Temperature:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
59#
 Author| Post time 2023-6-25 11:54:21 | Show all posts
4-wire Temperature:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1276

Posts

1356

Credits

金牌会员

Rank: 6Rank: 6

Credits
1356
60#
 Author| Post time 2023-7-1 08:11:25 | Show all posts
Step 2: Press [Temp] on the front panel. The following menu appears.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-18 12:37 GMT+8 , Processed in 0.176012 second(s), 21 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list