Electronic Engineer Discuss

 找回密码
 Register
楼主: WisdomAugust

Digital Multimeter HDM3055 Series Manual

[复制链接]

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-3-15 08:31:30 | 显示全部楼层
Errors in High Resistance Measurements

When you are measuring large resistances, significant errors can occur due to insulation resistance
and surface cleanliness. You should take the necessary precautions to maintain a "clean" high–resistance
system.
Test leads and fixtures are susceptible to leakage due to moisture absorption in insulating materials and
"dirty" surface films. Nylon and PVC are relatively poor insulators (109 ?) when compared to PTFE
insulators (1013 ?). Leakage from nylon or PVC insulators can easily contribute a 0.1% error when
measuring a 1 M? resistance in humid conditions.


回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-3-18 08:28:46 | 显示全部楼层
True RMS AC Measurements

True rms responding multimeters, like the HDM3000 Series, measure the "heating" potential of an
applied voltage. Power dissipated in a resistor is proportional to the square of an applied voltage,
independent of the waveshape of the signal. This multimeter accurately measures true rms voltage
or current, as long as the wave shape contains negligible energy above the meter’s effective bandwidth.

The HDM3000 Series uses the same techniques to measure true rms voltage and true rms current.

The effective AC voltage bandwidth is 300 kHz, while the effective AC current bandwidth is 10 kHz.


回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-1 15:20:55 | 显示全部楼层


本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?Register

x
回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-2 08:39:50 | 显示全部楼层
The DMM's AC voltage and AC current functions measure the AC-coupled true rms value. In this
DMM, the "heating value" of only the AC components of the input waveform are measured (dc is
rejected). As seen in the figure above; for sine waves, triangle waves, and square waves, the
AC–coupled and AC+DC values are equal, because these waveforms do not contain a DC offset.
However, for non–symmetrical waveforms (such as pulse trains) there is a DC voltage content,
which is rejected by Hantek’s AC–coupled true rms measurements. This can provide a significant
benefit.

An AC–coupled true rms measurement is desirable when you are measuring small AC signals in
the presence of large DC offsets. For example, this situation is common when measuring AC
ripple present on DC power supplies. There are situations, however, where you might want to
know the AC+DC true rms value.


回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-3 13:08:15 | 显示全部楼层
You can determine this value by combining results from DC and AC measurements, as shown
below:
file:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml5148/wps1.png
For the best AC noise rejection, you should perform the DC measurement using an integration time
of at least 10 power–line cycles (PLCs).


本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?Register

x
回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-7 11:38:57 | 显示全部楼层
True RMS Accuracy and High–Frequency Signal Content

A common misconception is that because an AC multimeter is true rms, its sine wave accuracy
specifications apply to all waveforms. Actually, the shape of the input signal dramatically affects
measurement accuracy for any multimeter, especially when that input signal contains high–
frequency the instrument’s bandwidth.

For example, consider a pulse train, one of the most challenging waveforms for a multimeter.
The pulse width of that waveform largely determines its high–frequency content. The frequency
spectrum of an individual pulse is determined by its Fourier Integral. The frequency spectrum
of the pulse train is the Fourier Series that samples along the Fourier Integral at multiples of
the input pulse repetition frequency (prf).

The figure below shows the Fourier Integral of two very different pulses: one of broad width
(200 ?s); the other narrow (6.7 ?s). The bandwidth of the ACV path in the DMM is 300 kHZ;
therefore, frequency content above 300 kHz is not measured.

Notice that the sin(πfT)/πfT spectrum of the narrow pulse significantly exceeds the effective
bandwidth of the instrument. The net result is a less accurate measurement of the narrow,
high–frequency pulse.


回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-8 08:55:51 | 显示全部楼层
In contrast, the frequency spectrum of the broad pulse has fallen off significantly below the multimeter’s
300 kHz (approximately) bandwidth, so measurements of this pulse are more accurate.

Reducing the prf increases the density of lines in the Fourier spectrum, and increases the portion of
the input signal’s spectral energy within the multimeter’s bandwidth, which improves accuracy.

In summary, error in rms measurements arise when there is significant input signal energy at
frequencies above the multimeter’s bandwidth.



本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?Register

x
回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-10 08:51:19 | 显示全部楼层
Estimating High–Frequency (Out–of–Band) Error
A common way to describe signal waveshapes is to refer to their "Crest Factor". Crest factor is the
ratio of the peak value to rms value of a waveform. For a pulse train, for example, the crest factor
is approximately equal to the square root of the inverse of the duty cycle.



本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?Register

x
回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-11 08:45:52 | 显示全部楼层
Notice that crest factor is a composite parameter, dependent upon the pulse width and repetition
frequency; crest factor alone is not enough to characterize the frequency content of a signal.

Traditionally, DMMs include a crest factor derating table that applies at all frequencies. The
measurement algorithm used in the DMMs is not inherently sensitive to crest factor, so no such
derating is necessary. With this multimeter, as discussed in the previous section, the focal issue
is high–frequency signal content which exceeds the multimeter’s bandwidth.

For periodic signals, the combination of crest factor and repetition rate can suggest the amount of
high– frequency content and associated measurement error. The first zero crossing of a simple
pulse occurs at f1 = 1/tp .

This gives an immediate impression of the high-frequency content by identifying where this crossing
occurs as a function of crest factor: f1=(CF2)(prf).


回复 支持 反对

使用道具 举报

58

主题

1293

帖子

1373

积分

金牌会员

Rank: 6Rank: 6

积分
1373
 楼主| 发表于 2024-4-12 08:09:54 | 显示全部楼层
The following table shows the typical error for various pulse waveforms as a function of input pulse
frequency.



本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?Register

x
回复 支持 反对

使用道具 举报

您需要登录后才可以回帖 登录 | Register

本版积分规则

Dark room|Mobile|Archiver|Electronic Engineer Discuss

GMT+8, 2024-6-26 09:36 , Processed in 0.253455 second(s), 20 queries .

Powered by Discuz! X3.2

Copyright © 2001-2021, Tencent Cloud.

快速回复 返回顶部 返回列表