Title: Problem hantek 6022be
Hello
I have a problem with the 6022be oscilloscope. when I connect the probe to a voltage below the 5v the right law , while for any voltage over 5v I find a value of 5,18v voltage . for example if you connect the probe to a 12v battery 5,18v reads me . how do I fix ?



Author: amy    Time: 2016-5-30 10:45



Dir sir,
The max. measure voltage is -5V~+5V without any attenuation. The max. measure voltage is -50V~+50V with 10:1 attenuation.
When you want to measure the voltage over 5V, please change the attenuation to 10:1.

Author: Galfo    Time: 2016-5-30 17:49

thanks for the reply .
In the channel because it says 35V max ? I can read voltages above 35V with attenuation commissioning probe to x10 ?

Author: amy    Time: 2016-6-1 08:42
Galfo replied at 2016-5-30 17:49
thanks for the reply .
In the channel because it says 35V max ? I can read voltages above 35V with  ...

The input protection is 35V. If the input voltage is over 35V, the device will be damaged.