AC Voltage
This section describes how to configure AC voltage measurements from the front panel.
NOTE Default delays are selected to give correct first readings for most measurements. For the most
accurate measurements, the input blocking RC time constant must settle to 1/50 of the AC signal level.
Signals greater than 300 V (rms) or 1 A (rms) will cause self-heating in signal-conditioning components.
These errors are included in the instrument specifications. Internal temperature changes due to
self-heating may cause additional error on other functions or ranges. The additional error will generally
dissipate within a few minutes.

For example, consider a 100 mVAC signal with a 10 VDC bias. The 10 VDC bias should be settled to
1/50 of 100 mVAC, or 2 mVDC. The corresponding settling time can be calculated using the blocking
RC time constant of 0.22 s as follows:

settling time = ln(bias/settled value)* 0.22 s
settling time = ln(10 VDC / 2 mVDC)* 0.22 s
settling time = ln(5000)* 0.22 s = 1.9 s


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
35#
 Author| Post time 2023-4-8 08:38:28 | Show the author posts only
DCV Ratio

The DCV Ratio key enables or disables DCV Ratio measurement. Note that the Auto Zero softkey
disappears when you enable DCV Ratio measurements. This is because autozero cannot be disabled
during DCV Ratio.
The ratio is the voltage on the Input terminals divided by the reference voltage. The reference voltage
is the difference of two separate measurements. These measurements are the DC voltages from the
HI Sense terminal to the LO Input terminal and from the LO Sense terminal to the LO Input terminal.
These two measurements must be within the range of ±12 VDC. The reference voltage is always
autoranged, and the range used for both will be based on the larger result of these two measurements.
Configure DCV Ratio measurements as shown:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
34#
 Author| Post time 2023-4-7 08:13:38 | Show the author posts only
Step 6:

Specify the input impedance to the test leads (Input Z). This specifies the measurement
terminal input impedance, which is either Auto or 10 MΩ. The Auto mode selects high impedance
(HighZ) for the 100 mV, 1 V, and 10 V ranges, and 10 MΩ for the 100 V and 1000 V ranges. In most
situations, 10 MΩ is high enough to not load most circuits, but low enough to make readings stable for
high-impedance circuits. It also leads to readings with less noise than the HighZ option, which is
included for situations where the 10 MΩ load is significant.


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
33#
 Author| Post time 2023-4-6 08:27:09 | Show the author posts only
Step 5:

Auto Zero: Autozero provides the most accurate measurements, but requires additional
time to perform the zero measurement. With autozero enabled (On), the DMM internally measures
the offset following each measurement. It then subtracts that measurement from the preceding
reading. This prevents offset voltages present on the DMM’s input circuitry from affecting measurement
accuracy. With autozero disabled (Off), the DMM measures the offset once and subtracts the offset
from all subsequent measurements. The DMM takes a new offset measurement each time you change
the function, range, or integration time. ( There is no autozero setting for 4-wire measurements.)


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
32#
 Author| Post time 2023-4-4 08:13:44 | Show the author posts only
Step 4:Press Range to select a range for the measurement. You can also use the [+], [-], and
[Range] keys on the front panel to select the range. (Auto (autorange) automatically selects
the range for the measurement based on the input. Autoranging is convenient, but it results in
slower measurements than using a manual range. Autoranging goes up a range at 120% of the
present range, and down a range below 10% of the present range.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
31#
 Author| Post time 2023-4-3 08:34:17 | Show the author posts only
Step 2: Press [DCV] on the front panel.
Step 3:
Press Aperture and choose the number of power-line cycles (PLCs) to use for the measurement.
1, 10, and 100 PLC provide normal mode (line frequency noise) rejection. Selecting 100 PLC provides
the best noise rejection and resolution, but the slowest measurements:




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
30#
 Author| Post time 2023-3-10 09:09:00 | Show the author posts only
Measurements

DC Voltage
This section describes how to configure DC voltage measurements from the front panel, including
DCV ratio measurements.
Step 1: Configure the test leads as shown.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
29#
 Author| Post time 2023-3-9 13:24:51 | Show the author posts only


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
28#
 Author| Post time 2023-3-8 08:38:25 | Show the author posts only


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
27#
 Author| Post time 2023-3-7 08:31:40 | Show the author posts only

Software and Driver of RS 4CH Oscilloscope



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-7-5 18:04 GMT+8 , Processed in 0.162442 second(s), 20 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list