Noise Rejection

Rejecting Power–Line Noise Voltages

A desirable characteristic of integrating analog–to–digital (A/D) converters is their ability to reject
power– line related noise present with DC input signals. This is called normal mode noise rejection,
or NMR. The multimeter achieves NMR by measuring the average DC input by "integrating" it over a
fixed period. If you set the integration time to a whole number of power line cycles (PLCs), these
errors (and their harmonics) will average out to approximately zero.

The multimeter provides three integration selections (1, 10, and 100 PLCs) that achieve NMR. The
multimeter measures the power–line frequency (50 Hz or 60 Hz), and then determines the corresponding
integration time. For a complete listing of NMR, approximate added rms noise, reading rate, and resolution
for each integration setting, see the Performance vs. Integration Time table.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
171#
 Author| Post time 2024-3-1 10:12:55 | Show the author posts only
Loading Errors (DC Voltage)

Measurement loading errors occur when the resistance of the DUT is an appreciable percentage of
the multimeter's input resistance, as shown below.


To reduce the effects of loading errors and to minimize noise pickup, set the multimeter's input
resistance to >10 G? (high-impedance) for the 100 mVDC, 1 VDC, and 10 VDC ranges. The
input resistance is maintained at 10 M? for the 100 VDC and 1000 VDC ranges.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
170#
 Author| Post time 2024-2-26 08:28:11 | Show the author posts only


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
169#
 Author| Post time 2024-2-24 13:10:51 | Show the author posts only
Measurement Tutorial

Follow the suggestions in the sections below to achieve the greatest accuracy with Hantek HDM3000
Series multimeters.

Measurement Considerations
To achieve the greatest accuracy with the Hantek HDM3000 Series multimeters, you must eliminate
potential measurement errors. This chapter describes common errors and suggests ways to avoid them.
DC Measurement Considerations

Thermal EMF Errors
Thermoelectric voltages are the most common error source in low–level DC voltage measurements.
Thermoelectric voltages are generated by circuit connections using dissimilar metals at different temperatures.
Each metal–to–metal junction forms a thermocouple that generates a voltage proportional to the junction
temperature, as shown in the table below. You should minimize thermocouple voltages and temperature
variations in low–level voltage measurements. The best connections use copper–to–copper crimped
connections, as the DMM’s input terminals are a copper alloy.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
168#
 Author| Post time 2024-2-23 08:45:32 | Show the author posts only
Display Options

Display Options configures the display.


You can enable or disable the display, adjust the brightness (1 to 100%). If you turn off the display,
press any front panel key to turn it on again.

Date / Time

Date / Time sets the instrument's real-time clock, which always uses a 24-hour format (00:00:00 to 23:59:59).
There is no automatic setting of the date and time, such as to adjust for daylight savings time.
Use the front panel arrow keys to set the year, month, day, hour, and minute.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
167#
 Author| Post time 2024-2-22 08:35:37 | Show the author posts only
Disables or enables the click heard when a front panel key or softkey is pressed.

Also enables or disables the audible tone (Beeper On or Off) associated with the following features:
l Limits – a measurement limit is exceeded (if limits are enabled)
l Probe Hold – the measured signal records a stable reading
l Diode – the forward bias voltage is between 0.3 and 0.8 V
l Continuity – a short is measured (less than or equal to 10 Ω)
l Error - an error is generated from the front panel or remote interface

This non-volatile setting appears in several different menus on the front panel. Turning the beeper on
or off in one menu affects all of the other menus and functions. For example, turning the beeper off for
probe hold, also turns the beeper off for limits, diode, continuity, and errors.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
166#
 Author| Post time 2024-2-21 08:31:16 | Show the author posts only
User Settings

User Settings specifies user preferences that control how you interact with the instrument. These
settings are stored in non-volatile memory and can be saved in a preferences (.prf) file.


Sounds



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
165#
 Author| Post time 2024-2-20 08:37:18 | Show the author posts only
Utility Menu - System Setup


System Setup configures user preferences and sets the date and time.




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
164#
 Author| Post time 2024-2-19 08:09:47 | Show the author posts only
Calibrate
Calibrate accesses the instrument calibration procedure. See Calibration for details.


Firmware Update
Firmware Update updates the instrument firmware to a new version. See Firmware Update
for details.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
163#
 Author| Post time 2024-2-7 08:16:17 | Show the author posts only
Utility Menu - Test/Admin


Test/Admin provides access to self-test, calibration, and firmware update function:


Self-Test
Self-Test verifies proper instrument operation. See Self-Test Procedures for details, and always safely
disconnect inputs to the DMM terminals before running the Self Test.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-29 09:05 GMT+8 , Processed in 0.201994 second(s), 20 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list