Introduction to the Instrument

file:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps1.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps2.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps3.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps4.pngFront Panel Review




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
Reply

Use magic Report

58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
223#
 Author| Post time 29 min. ago | Show the author posts only
Electrostatic Discharge (ESD) Precautions

Almost all electrical components can be damaged by electrostatic discharge (ESD) during handling.
Component damage can occur at electrostatic discharge voltages as low as 50 V.
The following guidelines will help prevent ESD damage during service operations:

l Disassemble instruments only in a static-free work area.
l Use a conductive work area to reduce static charges.
l Use a conductive wrist strap to reduce static charge accumulation.
l Minimize handling.
l Keep replacement parts in original static-free packaging.
l Remove all plastic, foam, vinyl, paper, and other static-generating materials from the immediate work
area.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
222#
 Author| Post time The day before yesterday 08:17 | Show the author posts only
Service and Repair

This section contains basic instrument service information.

Cleaning

To prevent electrical shock, disconnect the instrument from AC mains power and disconnect all test
leads before cleaning. Clean the outside of the instrument using a soft, lint-free, cloth slightly dampened
with water.
Do not use detergent or solvents.
Do not attempt to clean internally.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
221#
 Author| Post time 7 day(s) ago | Show the author posts only
When you change any of these settings, the sample interval (sample timer in continuous mode) is
increased to be greater than the calculated measurement time. In the continuous mode, attempting
to reduce the sample interval below the calculated value results in an error message. You must then
choose among the various ways you can achieve a smaller measurement time to achieve the smaller
sample interval, such as decreasing the NPLC setting.
Another way to put it is that for continuous mode, sample timer/interval is always controlled by the other
measurement settings.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
220#
 Author| Post time 2024-6-25 09:12:49 | Show the author posts only
How Sample Rate/Interval is Determined

The DMM always attempts to make the sample interval greater than the expected time required
to take the measurements. A number of settings go into the calculation of the minimum allowable
sample interval. These settings include the measurement function, NPLC, aperture, autorange,
autozero, offset compensation, AC filter, TC open check and gate time.

For example, when autorange is on, the assumption is that no more than one range change will
be required. If more than one level of change occurs, the measurement may be delayed and an
error will be issued.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
219#
 Author| Post time 2024-6-24 08:13:17 | Show the author posts only
Unnecessary Signal Errors

Both 3A and 10A terminals are available for AC and DC current measurements. If signals are applied
to terminals not being used for the current measurement, measurement errors may occur. The
unused terminals are protected but the additional signals may interfere with current measurements.
For example, applying inputs to the 3A terminals while making measurements on the 10A terminals
will typically cause errors.

Unnecessary signals applied to the Hi and Lo Sense terminals can also cause errors. AC or DC
voltages above 15 volts peak on the sense terminals are likely to cause measurement errors.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
218#
 Author| Post time 2024-6-21 09:33:27 | Show the author posts only
Leakage Current Errors

The DMM's input capacitance will "charge up" due to input bias currents when the terminals are
open–circuited (if the input resistance is >10 G?). The DMM's measuring circuitry exhibits
approximately 30pA of input bias current for ambient temperatures from 0 to 30 °C. Bias
current doubles for every 8 °C change in ambient temperature above 30 °C. This current
generates small voltage offsets dependent upon the source resistance of the DUT. This effect
becomes evident for a source resistance of greater than 100 k?, or
when the DMM's operating temperature is significantly greater than 30 °C.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
217#
 Author| Post time 2024-6-20 08:54:29 | Show the author posts only
Common Mode Errors

Errors are generated when the multimeter's input LO terminal is driven with an AC voltage relative
to earth. The most common situation where unnecessary common mode voltages are created is
when the output of an AC calibrator is connected to the multimeter "backwards." Ideally, a multimeter
reads the same regardless of how the source is connected. Both source and multimeter effects can
degrade this ideal situation. Because of the capacitance between the input LO terminal and earth
(approximately 200 pF), the source will experience different loading depending on how the input is
applied. The magnitude of the error is dependent upon the source's response to this loading.
The DMM's measurement circuitry, while extensively shielded, responds differently in the backward
input case due to slight differences in stray capacitance to earth. The DMM's errors are greatest for
high– voltage, high–frequency inputs. Typically, the DMM exhibits about 0.06% additional error for a
100 V, 100 kHz reverse input. You can use the grounding techniques described for DC common mode
problems to minimize AC common mode voltages.


58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
216#
 Author| Post time 2024-6-19 09:30:39 | Show the author posts only
Edited by WisdomAugust at 2024-6-20 08:52

Most extraneous noise is not correlated with the input signal. You can determine the error as shown
below.



Correlated noise, while rare, is especially detrimental because it always adds directly to the input signal.
Measuring a low–level signal with the same frequency as the local power line is a common situation that
is prone to this error.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1295

Posts

1375

Credits

金牌会员

Rank: 6Rank: 6

Credits
1375
215#
 Author| Post time 2024-6-18 08:08:53 | Show the author posts only
Connect the DMM and the AC source to the same electrical outlet whenever possible.
You should also minimize the area of any ground loops that cannot be avoided. A
high–impedance source is more susceptible to noise pickup than a low–impedance
source. You can reduce the high–frequency impedance of a source by placing a
capacitor in parallel with the DMM's input terminals. You may have to experiment
to determine the correct capacitor for your application.