For AC voltage, additional settling time may be required when the DC level varies from sample to
sample. The default sample delays allow for a DC level change of 3% of range for all filters. If the
DC level change exceeds these levels, additional settling time is required. The multimeter's DC
blocking circuitry has a settling time constant of 0.2 seconds. This settling time only affects
measurement accuracy when DC offset levels vary from sample to sample. If maximum
measurement speed is desired, you may want to add an external DC blocking circuit for circuitry
with significant DC voltages present. This circuit can be as simple as a resistor and a capacitor.


For AC current, additional settling time is not required when the DC level varies sample to sample.



58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
201#
 Author| Post time 2024-5-14 08:41:30 | Show the author posts only


In applications where sample–to–sample levels vary widely, but the DC offset level does not change,
the MEDIUM filter settles at 2 to 4 readings per second (depending upon the lowest frequency
component in the waveform) as shown in the following table:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
200#
 Author| Post time 2024-5-13 08:15:29 | Show the author posts only
Edited by WisdomAugust at 2024-5-14 08:37


With a few precautions, you can perform AC measurements at speeds up to 500 readings
per second.

Use manual ranging to eliminate auto-ranging delays. By setting the trigger delay to 0, the
FAST, MEDIUM, and SLOW filters allow up to 500, 150, and 50 readings per second, but with
reduced accuracy because the filter may not fully settle. If the sample–to–sample levels are
similar, little settling time is required for each new reading. Under this specialized condition,
the MEDIUM filter provides reduced accuracy results at 20 readings per second, and the FAST
filter provides reduced accuracy results at 200 readings per second.



58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
199#
 Author| Post time 2024-5-10 08:26:15 | Show the author posts only
Making High–Speed AC Measurements

The multimeter's AC voltage and AC current functions implement three low–frequency filters.
These filters allow you to trade–off minimum measured frequency for faster reading speed.
The FAST filter settles in 0.025 seconds, and is useful for frequencies above 200 Hz. The
MEDIUM filter settles in 0.625 seconds for voltage and 0.25 seconds for current, and is useful
for measurements above 20 Hz. The SLOW filter settles in 2.5 seconds for voltage and 1.66
seconds for current, and is useful for frequencies above 3 Hz.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
198#
 Author| Post time 2024-5-9 13:28:47 | Show the author posts only
NULL Reading

The DMM allows a separate null setting to be saved for the temperature function. When making
null measurements, each reading is the difference between a stored null value and the input signal.
One application of NULL is to increase accuracy of two-wire resistance measurements by first nulling
the closed–circuit test lead resistance.

Autozero On/Off

Enabling the autozero feature (ON) provides greater accuracy; however, the additional measurement
(of zero) reduces the reading speed.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
197#
 Author| Post time 2024-5-8 13:09:50 | Show the author posts only
Important: Use only a 5 k? 44007-type thermistor. This type thermistor has constants of A = 1.285e-3,
B = 2.362e-4, C = 9.285e-8. Using an incorrect type of thermistor can result in errors greater than 20 °C
for a temperature being measured of 100 °C.
2-Wire vs. 4-Wire Measurements

As with resistance measurements, 4-wire temperature measurements are more accurate, because errors
due to lead wire resistance are completely eliminated. Alternatively, you can use the multimeter’s Null function
to remove the test lead resistance from the measurement (see NULL Reading below).


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
196#
 Author| Post time 2024-5-7 08:23:00 | Show the author posts only
Thermistor Requirements

The DMM converts the measured thermistor resistance to temperature using the Steinhart-Hart
thermistor equation:
1?T = A + B (Ln(R)) + C (Ln(R))3
Where:
A, B, and C are constants provided by the thermistor manufacturer and derived from three
temperature test points.
R = Thermistor resistance in Ω.
T = Temperature in degrees K.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
195#
 Author| Post time 2024-4-20 08:28:03 | Show the author posts only
Temperature Measurements

Temperature measurements require a temperature transducer probe. The supported probes are
2-wire and 4-wire RTDs, 2-wire and 4-wire thermistors.
Probe Type Choice
RTD's provide very accurate, highly linear relationships between resistance and temperature,
over a range of roughly –200 to 500 °C. There is very little conversion complexity for an RTD
because it is so intrinsically linear. The multimeter provides measurement for the IEC751 standard
RTD, which has a sensitivity of 0.385%/°C.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
194#
 Author| Post time 2024-4-19 08:06:10 | Show the author posts only
Avoid applying signals to unused current input terminals

If signal inputs are applied to terminals not needed for the current measurement, measurement
errors may occur. The unused terminals are still protected but the un-needed signals may interfere
with current measurement. For example, applying inputs to the 3A terminals while making
measurements on the 10A terminals will typically cause errors.

The Hi and Lo sense terminals are not used for many measurements. Applying signals here when
not needed can also cause errors. AC or DC voltages above 15 volts peak on the un-needed sense
terminals are likely to cause measurement errors. If unexpected errors are occurring, signals on the
un-needed terminals is an area to check.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
193#
 Author| Post time 2024-4-18 11:42:54 | Show the author posts only
DC Current

When you connect the multimeter in series with a test circuit to measure current, a measurement
error is introduced. The error is caused by the multimeter's series burden voltage. A voltage is
developed across the wiring resistance and current shunt resistance of the multimeter, as shown
below.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-26 23:40 GMT+8 , Processed in 0.244685 second(s), 20 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list