amy replied at 2016-12-5 17:19
Peak: The scope of the oscilloscope finds the maximum and minimum input signal in each sampling interval and use these values to display the waveform.
Well, as I understood from the waveform of the two curves is normal. However, this is not the full answer to my question. In particular, I still do not understand, why then the installation scan time of 80 ms or slower and established Data Acquisition mode: Normal waveform drawn single curve:
and for manual mode Data Acquisition mode: Peak Detect waveform starts to be drawn double curve:
I do not understand why when you install the scan time of 80 ms or slower waveform in any case not drawn double curve? In the subject "DSO5000 Series Quick Start Guide" (http://www.eediscuss.com/forum.p ... tra=page%3D1&page=6) in the section "4.1.2 Slow Scan Mode" on what was said by others following:
In the Slow Scan mode, peak detect acquisition is used so that no data is missed (even the though the Acquire menu may show a different acquisition mode setting).
That is, as far as I understand, when you install the scan time of 80 ms or slower, regardless of how Data Acquisition mode has been enabled manually, the oscilloscope should automatically switch to the Data Acquisition mode mode: Peak Detect, and draw the waveform double curves, however, this does not happen ... Why this happens? Maybe I have something here do not mean right?
|
|