I don't think it's relevant, happens to everyone, but here you have:
DSO2D10
CN2101029001641
I'm running latest SW.
As I said, the zoom worked correctly in older firmware version up to 2021.01.05, but in these old software versions there were other issues, like poor trigger detection on small pulses.
Newer firmware fixed some bugs, but introduced others, like the zoom problem.
Additionally, I recently found another important bug when trying to capture a single happenning pulse.
Steps to reproduce:
- Set trigger in NORMAL mode, any edge mode, any time division.
- Run a small pulse. The trigger captures and stores the wave because it's in NORMAL mode (Doesn't refresh until next trigger happens)
- I push STOP to stop the sampling and keep the wave in the screen. The indicator becomes red.
- But when removing the probe or conecting anything else, the scope captures another edge and clears the stored wave.
Setting the scope into STOP (Red), does exactly the same thing as Single SEQ, triggers only once, but shouldn't trigger at all.This seems to happen on both old and new software versions.
Also, when pressing run/stop button, often the indicator does strange things, should switch between Red and Green, but very often it goes Orange (Screen shows READY) or Off (Screen shows STOP), seems like something is not updating properly?
|
|