Loading Errors (DC Voltage)

Measurement loading errors occur when the resistance of the DUT is an appreciable percentage of
the multimeter's input resistance, as shown below.


To reduce the effects of loading errors and to minimize noise pickup, set the multimeter's input
resistance to >10 G? (high-impedance) for the 100 mVDC, 1 VDC, and 10 VDC ranges. The
input resistance is maintained at 10 M? for the 100 VDC and 1000 VDC ranges.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
172#
 Author| Post time 2024-3-4 14:51:01 | Show the author posts only
Noise Rejection

Rejecting Power–Line Noise Voltages

A desirable characteristic of integrating analog–to–digital (A/D) converters is their ability to reject
power– line related noise present with DC input signals. This is called normal mode noise rejection,
or NMR. The multimeter achieves NMR by measuring the average DC input by "integrating" it over a
fixed period. If you set the integration time to a whole number of power line cycles (PLCs), these
errors (and their harmonics) will average out to approximately zero.

The multimeter provides three integration selections (1, 10, and 100 PLCs) that achieve NMR. The
multimeter measures the power–line frequency (50 Hz or 60 Hz), and then determines the corresponding
integration time. For a complete listing of NMR, approximate added rms noise, reading rate, and resolution
for each integration setting, see the Performance vs. Integration Time table.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
173#
 Author| Post time 2024-3-6 08:10:56 | Show the author posts only
Common Mode Rejection (CMR)

Ideally, a multimeter is completely isolated from earth–referenced circuits. However, there is finite
resistance between the multimeter's input LO terminal and earth ground, as shown below. This can
cause errors when measuring low voltages which are floating relative to earth ground.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
174#
 Author| Post time 2024-3-7 08:14:22 | Show the author posts only
Noise Caused by Magnetic Loops

When making measurements near magnetic fields, avoid inducing voltages in the measurement
connections. You should be especially careful when working near conductors carrying large currents.
Use twisted–pair connections to the multimeter to reduce the noise pickup loop area, or dress the
test leads as close together as possible. Loose or vibrating test leads will also induce error voltages.
Tie down test leads
securely when operating near magnetic fields. Whenever possible, use magnetic shielding materials
or increased distance from magnetic sources.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
175#
 Author| Post time 2024-3-8 08:04:05 | Show the author posts only
Noise Caused by Ground Loops

When measuring voltages in circuits where the multimeter and the DUT are referenced to a common
earth ground, a "ground loop" is formed. As shown below, any voltage difference between the two
ground reference points (Vground) causes current to flow through the measurement leads. This
causes noise and offset voltage (usually power–line related), which are added to the measured voltage.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
176#
 Author| Post time 2024-3-9 08:13:20 | Show the author posts only


The best way to eliminate ground loops is to isolate the multimeter from earth by not grounding
the input terminals. If the multimeter must be earth–referenced, connect it and the DUT to one
common ground point. Also connect the multimeter and DUT to the same electrical outlet
whenever possible.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
177#
 Author| Post time 2024-3-11 08:09:32 | Show the author posts only
Resistance Measurement Considerations

The multimeter offers two resistance measurements: 2-wire and 4-wire ohms. For both methods,
the test current flows from the input HI terminal, through the resistor being measured. For 2-wire
ohms, the voltage drop across the resistor being measured is sensed internal to the multimeter.
Therefore, test lead resistance is also measured. For 4-wire ohms, separate "sense" connections
are required. Because no current flows in the sense leads, the resistance in these leads does not
give a measurement error.

The errors mentioned earlier in this chapter for DC voltage measurements also apply to resistance
measurements. Additional error sources unique to resistance measurements are discussed below.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
178#
 Author| Post time 2024-3-12 08:21:14 | Show the author posts only
The 4-wire ohms method provides the most accurate way to measure small resistances because
it reduces test lead and contact resistances. This is often used in automated test applications
where resistive and/or long cable, numerous connections, or switches exist between the multimeter
and the DUT. The recommended connections for 4-wire ohms measurements are shown below.




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
179#
 Author| Post time 2024-3-13 09:55:50 | Show the author posts only
Removing Test Lead Resistance Errors

To eliminate offset errors associated with test lead resistance in 2-wire ohms measurements,
follow these steps:

Short the test lead ends together and read the displayed test lead resistance.
Press Null. The multimeter will store the test lead resistance as the 2-wire ohms null value,
and subtract that value from subsequent measurements.
See also "Null Measurements."


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
180#
 Author| Post time 2024-3-14 08:12:45 | Show the author posts only
Minimizing Power Dissipation Effects

When measuring resistors designed for temperature measurements (or other resistive devices
with large temperature coefficients), be aware that the multimeter will dissipate some power in
the DUT. The following table shows several examples.


If power dissipation is a problem, you should select a higher fixed range (all multimeter models).


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-26 14:13 GMT+8 , Processed in 0.171581 second(s), 21 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list