In applications where sample–to–sample levels vary widely, but the DC offset level does not change,
the MEDIUM filter settles at 2 to 4 readings per second (depending upon the lowest frequency
component in the waveform) as shown in the following table:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
202#
 Author| Post time 2024-5-15 08:37:28 | Show the author posts only
Edited by WisdomAugust at 2024-5-16 08:10

For AC voltage, additional settling time may be required when the DC level varies from sample to
sample. The default sample delays allow for a DC level change of 3% of range for all filters. If the
DC level change exceeds these levels, additional settling time is required. The multimeter's DC
blocking circuitry has a settling time constant of 0.2 seconds. This settling time only affects
measurement accuracy when DC offset levels vary from sample to sample. If maximum
measurement speed is desired, you may want to add an external DC blocking circuit for circuitry
with significant DC voltages present. This circuit can be as simple as a resistor and a capacitor.


For AC current, additional settling time is not required when the DC level varies sample to sample.



58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
203#
 Author| Post time 2024-5-16 08:13:34 | Show the author posts only
Making High–Speed DC and Resistance Measurements

To make the fastest (but least accurate) DC or resistance measurements:

l Set the integration time (NPLC or aperture) to minimum
l Select a fixed range (autorange off)
l Disable autozero
l Disable offset compensation (resistance measurements)

Refer to the particular measurement type in Measurements for more information on the above
functions.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
204#
 Author| Post time 2024-5-17 08:10:20 | Show the author posts only
Capacitance

The multimeter makes capacitance measurements by applying a known current to charge the
capacitance and then a resistance to discharge as shown below:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
205#
 Author| Post time 2024-5-20 11:41:45 | Show the author posts only
Edited by WisdomAugust at 2024-5-21 08:10

An illustration of the response curve while charging is shown below:




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
206#
 Author| Post time 2024-5-21 08:12:48 | Show the author posts only
Capacitance is calculated by measuring the change in voltage (DV) that occurs over a “short aperture”
time, (Dt). This measurement is repeated at two different times during the exponential rise that occurs.
An algorithm takes the data from these four points, and by linearizing that exponential rise over these
“short apertures”, accurately calculates the capacitance value.
The measurement cycle consists of two parts: a charge phase (shown in the graph) and a discharge phase.
The time–constant during the discharge phase is longer, due to a 100 kΩ protective resistor in the measurement
path. This time–constant plays an important role in the resultant reading rate (measurement time). The
incremental times (or “sample times”) as well as the width of the “short apertures”, vary by range, in order
to minimize noise and increase reading accuracy.

For the best accuracy, take a zero null measurement with open probes, to null out the test lead capacitance,
before connecting the probes across the capacitor to be measured (see Capacitance Measurements for details).


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
207#
 Author| Post time 2024-5-22 08:44:08 | Show the author posts only
Capacitance Measurement Considerations

Capacitors that have a high dissipation factor or other non-ideal characteristics will affect capacitance
measurements. Capacitors with high dissipation factors may exhibit a variance between the measured
value using the multimeter versus the single frequency method of some other LCR meters. The single
frequency method will also see more variation at different frequencies. For example, some inexpensive
capacitance substitution boxes, when measured with the multimeter, are almost 5% different compared
to the same capacitance measured with the single frequency method of an LCR meter. The LCR meter
will also show different values at different frequencies.

Capacitors with long time constants (dielectric absorption) will result in slow measurement settling time,
and will take a number of seconds to stabilize. You may see this when first connecting a capacitor or
when the delay time to make a measurement is varied. A high quality film capacitor typically shows the
least of this and an electrolytic capacitor the most, with ceramic capacitors typically in between.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
208#
 Author| Post time 2024-5-23 08:52:37 | Show the author posts only
Other Sources of Measurement Error
Settling Time Effects

The multimeter can insert automatic measurement settling delays. These delays are adequate for
resistance measurements with less than 200 pF of combined cable and device capacitance. This is
particularly important when measuring resistances above 100 k?. Settling due to RC time constant
effects can be quite long. Some precision resistors and multi–function calibrators use large parallel
capacitors (1000 pF to 0.1 ?F) with high resistor values to filter out noise currents injected by their
internal circuitry. Non–ideal capacitances in cables and other devices may have much longer settling
times than expected just by RC time constants due to dielectric absorption (soak) effects. Errors are
measured when settling after the initial connection and after a range change.


58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
209#
 Author| Post time 2024-6-11 08:12:03 | Show the author posts only
Loading Errors (AC volts)

In the AC voltage function, the input of the multimeter appears as a 1 M? resistance in parallel with
100 pF of capacitance. The cabling that you use to connect signals to the multimeter also adds
capacitance and loading. The table below shows the multimeter's approximate input resistance at
various frequencies.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1293

Posts

1373

Credits

金牌会员

Rank: 6Rank: 6

Credits
1373
210#
 Author| Post time 2024-6-12 15:22:43 | Show the author posts only
Edited by WisdomAugust at 2024-6-13 08:03

For low frequencies, the loading error is


At high frequencies, the additional loading error is




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-26 13:33 GMT+8 , Processed in 0.173947 second(s), 21 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list