When you change any of these settings, the sample interval (sample timer in continuous mode) is
increased to be greater than the calculated measurement time. In the continuous mode, attempting
to reduce the sample interval below the calculated value results in an error message. You must then
choose among the various ways you can achieve a smaller measurement time to achieve the smaller
sample interval, such as decreasing the NPLC setting.
Another way to put it is that for continuous mode, sample timer/interval is always controlled by the other
measurement settings.


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
222#
 Author| Post time 4 day(s) ago | Show the author posts only
Service and Repair

This section contains basic instrument service information.

Cleaning

To prevent electrical shock, disconnect the instrument from AC mains power and disconnect all test
leads before cleaning. Clean the outside of the instrument using a soft, lint-free, cloth slightly dampened
with water.
Do not use detergent or solvents.
Do not attempt to clean internally.


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
223#
 Author| Post time The day before yesterday 08:17 | Show the author posts only
Electrostatic Discharge (ESD) Precautions

Almost all electrical components can be damaged by electrostatic discharge (ESD) during handling.
Component damage can occur at electrostatic discharge voltages as low as 50 V.
The following guidelines will help prevent ESD damage during service operations:

l Disassemble instruments only in a static-free work area.
l Use a conductive work area to reduce static charges.
l Use a conductive wrist strap to reduce static charge accumulation.
l Minimize handling.
l Keep replacement parts in original static-free packaging.
l Remove all plastic, foam, vinyl, paper, and other static-generating materials from the immediate work
area.


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
224#
 Author| Post time Yesterday 08:16 | Show the author posts only
Troubleshooting

Before troubleshooting or repairing the instrument, make sure the failure is in the instrument rather
than any external connections. Also make sure that the instrument was accurately calibrated within
the last year (see Calibration Interval). The instrument’s circuits allow troubleshooting and repair with
basic test equipment.
CAUTION DO NOT swap the motherboard or the front panel board from one instrument to another.
These boards contain model number and serial number information that uniquely identifies a specific
unit, and boards that are mismatched to the instrument may result in problems with its performance
or warranty.


58

Threads

1297

Posts

1377

Credits

金牌会员

Rank: 6Rank: 6

Credits
1377
225#
 Author| Post time 4 hour(s) ago | Show the author posts only
Troubleshooting Procedure
1 Remove all remote I/O connections and front panel connections to the instrument. Verify that:
a. the AC mains power cord is securely connected to the instrument and plugged into a live outlet
b. the front panel Power On/Standby switch has been pushed
2 If the standby light below the power switch is not illuminated, re-verify the items above. If all this
is correct, check the AC mains power and then the internal line fuse.
3 Turn the instrument on. If the power-on sequence completes and the display appears to function
properly, verify that the instrument is in local mode, and run the self-test ([Utility] > Test/Admin >
Self-Test). If the display is illegible, replace the front panel board.
CAUTION  Ensure that all connections (front and rear) are removed when self-test is performed.
During self-test, errors may be induced by signals present on external wiring, such as long test
leads that can act as antennas.