Title: 6074BD - offset errors [Print this page]

Author: gf1    Time: 2019-3-10 00:52
Title: 6074BD - offset errors
Edited by gf1 at 2019-3-10 01:30

Dear Support,

on my 6074BD I notice significant offset errors.
Utility -> Calibration does not eliminate them.
The following image was captured immediately after running "Calibration":

[attach]2674[/attach]

As you can see, CH1 has an offset error of more than 1 div (and at 2mV/div, the error becomes even 2 div).
CH2 and CH4 have some offset error as well. Only CH3 happens to be almost OK (but still a bit off).

Btw, all inputs are terminated with 50 Ohm resistors - no signal present.


Turning the 20MHz bandwidth limit on or off also moves the trace a little but up or down.

The AC/DC switch seems not to affect the offset

After calibrating all 4 channels, I turned off CH2, CH3, CH4 and switched to a faster timebase of 500 and 200 ns/div. The offset of CH1 did change to than 5 div:

[attach]2675[/attach]

Sotware version 20160628
FPGA version D001
PCB version V1.06

[ I had installed the software from the official product page. Is there any newer versionavaiable? ]

gf1



Author: OASJ2YSeeEhBQo1    Time: 2019-3-10 07:21
Interesting, I am having exactly the same problem with some 6074BE scopes and was about to send an email to the support when I saw this thread.

[attach]2676[/attach]

In the manual, it is stated that all probes must be removed from the input before calibration (as it is normal in all oscilloscopes that I know off) but, in the software, in the Calibration windows, it says that you should short the inputs to ground and then run the calibration. So I decided to try that and apparently it solves the problem but after a while the offset problem returns.

My questions are:
1) Is this a hardware or software related problem? And how to fix it?
2) What is the correct procedure for calibration? The one described in the manual or the one in the software menu?

regards.



Author: Han    Time: 2019-3-11 08:45
Hantek will reply in 24 hours

Author: gf1    Time: 2019-3-12 03:19
Edited by gf1 at 2019-3-12 03:21

I do not see a statistically significant difference with open inputs and 50 Ohm terminated inputs. The resulting offsets vary anyway when I run the calibration multiple times in sequence. And after a calibration run, the residual offset is not invariant either, for each channel, but varies with the set of enabled channels, V/div, time/div and bandwidth limit setting.


[ It were certainly not the best idea, though, to keep the probe connected and to short the ground lead to the probe tip, since this "loop antenna" may pickup noise from the environment. ]

gf1





Author: OASJ2YSeeEhBQo1    Time: 2019-3-12 12:39
Does the 6074 scopes and the new Hantek2000 share a similar frontend or piece of firmware? I'm asking because I am observing a similar problem with my 2D42 and as gf1 noticed, I can clearly see the offset of trace or waveform changing as I change the scale setting or turn on/off a channel. Calibrating didn't help as well.

I will try to make a quick video later to better illustrate the issue.

Author: gf1    Time: 2019-3-13 07:25
Edited by gf1 at 2019-3-14 03:04

When I look at the PCB photos available in the internet, then V/div attenuators, ADC driver, channel multiplexer look quite different in the 2000 and 6000BD series. One common sub-circuit is obviously the dual-path input buffer amplifier, which uses a discrete FET source follower on the fast AC path, and an AD9510 on the DC path. This sub-circuit seems to be present - with small variations - in several scopes (IIRC also in some Rigol ones).

gf1



Author: OASJ2YSeeEhBQo1    Time: 2019-3-24 12:15
Edited by OASJ2YSeeEhBQo1 at 2019-3-24 22:55

I got in contact with Hantek by e-mail and they replied with one possible solution.

Try this: connect the black connector of the USB cable to your computer and the red one connect to an external power supply (like an USB charger) but it must be a good one, capable of supply 5V@2A then try the calibration again. For my case it helped and I can see an improvement although some offset still visible in the lower volt/div settings.

Apparently the problem is related to insufficient power being supplied to the device.



Author: gf1    Time: 2019-3-26 01:47
OASJ2YSeeEhBQo1 replied at 2019-3-24 12:15
I got in contact with Hantek by e-mail and they replied with one possible solution.

Try this: conne ...

Thank you, I will try...



Author: OASJ2YSeeEhBQo1    Time: 2019-3-26 14:36
I've being doing some more tests with a 6074BE and I think I may have found the problem, well at least one of them...

I am almost convinced that the main cause of this offset error is a software problem. I noticed that the initial position of the channel's cursor (the arrow on the left) on the screen influences the results of the calibration. Setting the cursors at the center of the screen before start the calibration yields the best results for me. I also noticed that changing the time/div to below 1us increases the offset. Changing the power supplied to the scope also has influence, so the thing is not so simple...

I am planing to do a more systematic test later and I'll report back with more details.



Author: amy    Time: 2019-3-27 10:55
Sir, yes, we found that there was a bug with the software. Oue engineer planned to finish the modification by the end of this week.
Please wait a while.

Author: OASJ2YSeeEhBQo1    Time: 2019-4-6 07:34
Dear Amy,

Please let me know when you have the software update ready with the fix, I'm needing that.

thx

Author: OASJ2YSeeEhBQo1    Time: 2019-4-9 08:21
Ok, so after many tests (I won't describe all them, it is not important) I am 100% sure that the offset problem is cause by a bug in the PC software and I found a temporary fix for it, at least until Hantek release an update.

The bug affects the version 2.2.4 of the software available for download on the product's web page. To fix the offset, install an older version (I've used Ver. 2.1.36, included in the CD) and run the calibration routine, done.

Result of the calibration using Ver. 2.2.4 of the software:

[attach]2778[/attach]

The same scope after re-running calibration using Ver. 2.1.36:
[attach]2779[/attach]

I want a beer




Author: amy    Time: 2019-4-9 15:18
Sorry to have a holiday, I'll urge the engineer to finish tomorrow.

Author: amy    Time: 2019-4-10 17:02
The software has been updated. Please download in this link: http://hantek.com/en/ProductDetail_2_10164.html

Author: gf1    Time: 2019-4-11 02:45
Thank you! Works much better now

Author: OASJ2YSeeEhBQo1    Time: 2019-4-11 10:29
Thx, the new software update (Version 2.2.5) apparently solved the offset problem.

Author: liziaq88    Time: 2019-4-27 12:33
I just got hantek 1008C, but can't download the software and the disc is empty. What should I do

Author: Han    Time: 2019-4-27 13:19
You can download from Hantek's website

Author: amy    Time: 2019-4-27 16:41
lateefomeiza88@gmail.com replied at 2019-4-27 12:33
I just got hantek 1008C, but can't download the software and the disc is empty. What should I do

Please download in this link: http://hantek.com/en/ProductDetail_13_13170.html

Author: liziaq88    Time: 2019-7-16 22:42
amy replied at 2019-4-27 16:41
Please download in this link: http://hantek.com/en/ProductDetail_13_13170.html

now i have the software,but whenever i connected the hantek 1008C device ,it keep displaying"SEARCHING DEVICE" and the device was never found.
i initially thought it was the windows ,then later try it with window 10 and lastly with window 8,the outcome was always the same.
please kindly help provide solution to this .

thanks

Author: Han    Time: 2019-7-17 13:29
Have you install the driver?