Edited by wjw at 2019-9-25 22:00
In Average/128/40k mode
I have a DSO4202c with software version 3.50.0(190325.0).
If I put it in Average mode, 128 average, memory depth 40k, it will reset after a few minutes. All the settings are lost, and it returns to operating with default settings.
I was trying this combination because when I look at a 21 MHz sinewave, it looks "jittery". I don't see this jittery distortion when I look at the same sinewave with other oscilloscopes.
Other settings
sec/div: 40.0 ns
Type: Real Time
volts/div: 5.00V
input: on Ch.1
While switching between CH1 & CH2
While switching between CH1 & CH2 to set up the scope, it will reset.
Procedure (after a reset has occurred):
Set SEC/DIV to 20.0ns
Press the CH2 button to turn on CH2
Set CH1 VOLT/DIV to 5.0, then CH2 to 5.0 with the knobs
Press the CH1 button to switch to the CH1 setup menu
Press the CH1 button to turn OFF CH1 (this is the first time)
Press the CH1 button again to turn CH1 ON (this is the first time)
Press the CH1 button again to turn OFF CH1 (this is the second time)
Press the CH1 button again to turn CH1 ON (this is the second time)
The scope will reset.
|
|