Electronic Engineer Discuss

Author: WisdomAugust
Print Prev. thread Next thread

DSO2000 Series SCPI Programmers Manual

[Copy link]

58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
21#
 Author| Post time 2021-8-10 09:40:34 | Show all posts
Edited by WisdomAugust at 2021-8-10 09:44

3. ACQuire Command Subsystem
Command List
l ACQuire: POINts
l ACQuire:TYPE
l ACQuire:SRATe?
l ACQuire:COUNt


3.1 ACQuire: POINts
Command
:ACQuire: POINts <value>
:ACQuire: POINts?
Description
Set or query the storage depth of the oscilloscope[size=10.5000pt].
Parameters
:<value> ::= [size=10.5000pt]memory depth value.
State
value                                
Memory depth
4000
4K
40000
40K
400000
400K
4000000
4M
8000000
8M
Return
The query returns the actual number of points (integer).
Example
:ACQuire: POINts 4000 /*Set memory depth to 4k*/
:ACQuire: POINts? /*The query returns[size=10.5000pt] 4000*/







58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
22#
 Author| Post time 2021-9-1 14:54:23 | Show all posts
Edited by WisdomAugust at 2021-9-1 14:56

3.2 ACQuire:TYPE
Command
:ACQuire:TYPE <value>
:ACQuire:TYPE?
Description
Set or query the acquisition method of oscilloscope samples
Parameters
:<value> ::= < NORMal | AVERage | PEAK | HRESolution>
State
?NORMal (normal): In this mode, the oscilloscope samples the signal at equal
time intervals to reconstruct the waveform. For most waveforms, using this
mode produces the best display results.
?AVERages: In this mode, the oscilloscope averages waveforms that have been 18 / 136
sampled multiple times to reduce random noise on the input signal and
improve vertical resolution. The number of averages can be set by the: ACQuire:
AVERages command. The higher the number of averages, the lower the noise
and the higher the vertical resolution, but the slower the response of the
displayed waveform to waveform changes.
?PEAK (Peak Detection): In this mode, the oscilloscope collects the maximum
and minimum values of the sampling interval signal to obtain the signal
envelope or narrow pulses that may be lost. Using this mode can avoid aliasing
of the signal, but the display noise is relatively large.
?HRESolution (high resolution): This mode uses an oversampling technique to
average the neighboring points of the sampled waveform to reduce random
noise on the input signal and produce a smoother waveform on the screen. It is
usually used when the sampling rate of the digitizer is higher than the storage
rate of the acquisition memory.
Return
The query returns NORM, AVERage, PEAK, or HRESolution.
Example
:ACQuire:TYPE AVERage[size=10.5000pt] /*Set the acquisition method to average*/
:ACQuire:TYPE?[size=10.5000pt] /*The query returns AVERage*/




58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
23#
 Author| Post time 2021-9-2 11:57:58 | Show all posts
Edited by WisdomAugust at 2021-9-2 12:00

3.3 ACQuire:SRATe?
Command
:ACQuire:SRATe?
Description
Query the current sampling rate.
State
The number of waveform points sampled per second.
Return
The query returns the sample rate as a real number.
Example
:ACQuire:SRATe?[size=10.5000pt] /*The query returns 1.25e+06*/


3.4 ACQuire:COUNt
Command
:ACQuire:COUNt <value>
:ACQuire:COUNt?
Description
Set or query the average number of times in the average acquisition mode
Parameters
:<value> ::= <4|8|16|32|64|128>
Return
The query returns current average times.
Example
:ACQuire:COUNt  64 /*Set the average number to 64*/
:ACQuire:COUNt? /*The query returns 64*/





58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
24#
 Author| Post time 2021-9-3 13:53:34 | Show all posts
4. TRIGger Command Subsystem
Command List
l TRIGger:FORCe
l TRIGger:MODE
l TRIGger:STATus?
l TRIGger:SWEep
l TRIGger:HOLDoff
l TRIGger:EDGe
l TRIGger: PULSe
l TRIGger:SLOPe
l TRIGger:TV
l TRIGger:TIMeout
l TRIGger:WINDOw
l TRIGger:INTERVAl
l TRIGger:UNDER_Am
l TRIGger:UART
l TRIGger:CAN
l TRIGger: LIN
l TRIGger:IIC
l TRIGger:SPI
l TRIGger: PATTern



58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
25#
 Author| Post time 2021-9-6 08:23:49 | Show all posts
Edited by WisdomAugust at 2021-9-6 08:26

4.1 TRIGger:FORCe
Command
:TRIGger:FORCe
Description
Select or query the force trigger.
State
Even if the trigger conditions are not met, this command will let the oscilloscope acquire waveforms
4.2TRIGger:MODE
Command
:TRIGger:MODE <mode>
:TRIGger:MODE?
Description
Select or query the trigger type.
Parameters
:<mode> ::= < EDGE | PULSe | TV | SLOPe | TIMeout | WINdow | PATTern | INTerval | UNDerthrow | UART | LIN | CAN | SPI | IIC>
Return
Return EDGE, PULSe, TV, SLOPe, TIMeout, WINdow, PATTern, INTerval, UNDerthrow, UART, LIN, CAN, SPI, IIC
Example
:TRIGger:MODE SLOPe[size=10.5000pt] /*Set the trigger type to slope trigger*/
:TRIGger:MODE? /*The query returns SLOPe*/
4.3TRIGger:STATus?
Command
:TRIGger:STATus?
Description
Query the current trigger status.
Return
Return TRIGed, NOTRIG



58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
26#
 Author| Post time 2021-9-7 10:13:02 | Show all posts
Edited by WisdomAugust at 2021-9-7 10:49

4.4 TRIGger:SWEep
Command
:TRIGger:SWEep <value>
:TRIGger:SWEep?
Description
Set or query the trigger mode.
Parameters
:<value> ::= < AUTO | NORMal | SINGle>
State
?AUTO: Automatic triggering, waveform display regardless of whether trigger
conditions are met.
?NORMal: Normal trigger, display the waveform when the trigger condition is
met, keep the original waveform display when the trigger condition is not met,
and wait for the next trigger.
?SINGle: Single trigger. The oscilloscope waits for the trigger, displays the
waveform when the trigger conditions are met, and then stops.
Return
The query returns AUTO, NORMal, or SINGle.
Example
:TRIGger:SWEep SINGle[size=10.5000pt] /*Select single trigger mode*/
:TRIGger:SWEep?[size=10.5000pt] /*The query returns SINGle*/
4.5 TRIGger:HOLDoff
Command
:TRIGger:HOLDoff <value>
:TRIGger:HOLDoff?
Description
Set or query the holdoff time.
Parameters
:<value> ::= Trigger holdoff time (in second)
State
Trigger holdoff can stably trigger complex waveforms (such as pulse series).
Holdoff time is the time the oscilloscope waits to re-enable the trigger circuit.
The oscilloscope will not trigger until the holdoff time is over.
When the trigger method is video trigger, timeout trigger, setup hold, UART,
LIN, CAN, IIC or SPI, this setting is not available.
Return
The query returns the trigger holdoff time in scientific notation.
Example
:TRIGger:HOLDoff 0.0000002[size=10.5000pt] /*Set trigger holdoff time to 200ns*/
:TRIGger:HOLDoff?[size=10.5000pt] /*The query returns 2.000000e-07*/




58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
27#
 Author| Post time 2021-9-8 11:04:39 | Show all posts
Edited by WisdomAugust at 2021-9-8 11:06

4.6 TRIGger:EDGe
Command List
l TRIGger:EDGe:SOURce
l TRIGger:EDGe:SLOPe
l TRIGger:EDGeEVel


4.6.1 TRIGger:EDGe:SOURce
Command
:TRIGger:EDGe:SOURce <source>
:TRIGger:EDGe:SOURce?
Description
Set or query the trigger source for edge trigger.
Parameters
:<source> ::= < CHANnel1 | CHANnel2 | CHANnel3 | CHANnel4 | EXT/10>
Return
The query returns CHANnel1, CHANnel2, CHANnel3, CHANnel4, EXT/10
Example
:TRIGger:EDGe:SOURce CHANnel1[size=10.5000pt] / * Set the trigger source to CH1 * /
:TRIGger:EDGe:SOURce?[size=10.5000pt] / * The query returns CHANnel1 * /




58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
28#
 Author| Post time 2021-9-9 09:45:08 | Show all posts
Edited by WisdomAugust at 2021-9-9 09:48

4.6.1 TRIGger:EDGe:SLOPe
Command
:TRIGger:EDGe:SLOPe <slope>
:TRIGger:EDGe:SLOPe?
Description
Set or query the type of edge trigger.
Parameters
:<slope> ::= < RISIng| FALLing | EITHer >
State
?RISIng: rising edge
?FALLing: falling edge
?EITHer: rising or falling edge
Return
The query returns RISIng, FALLing, EITHer
Example
:TRIGger:EDGe:SLOPe RISIng /*Set the edge type to rising edge*/
:TRIGger:EDGe:SLOPe? /*The query returns RISIng*/


4.6.2 TRIGger:EDGe: LEVel
Command
:TRIGger:EDGe: LEVel <level>
:TRIGger:EDGe: LEVel?
Description
Set or query the trigger level when the edge triggers.
Parameters
:<level> ::= trigger level ( unit is V)
Return
The query returns the trigger level value in scientific notation.
Example
:TRIGger:EDGe: LEVel 0.16 /*Set the trigger level to 160mV*/
:TRIGger:EDGe: LEVel? /*The query returns 1.600e-1*/







58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
29#
 Author| Post time 2021-10-8 16:31:03 | Show all posts
Edited by WisdomAugust at 2021-10-8 16:38

4.6.3 TRIGger:EDGe: LEVel
Command
:TRIGger:EDGe: LEVel <level>
:TRIGger:EDGe: LEVel?
Description
Set or query the trigger level when the edge triggers.
Parameters
:<level> ::= trigger level ( unit is V)
Return
The query returns the trigger level value in scientific notation.
Example
:TRIGger:EDGe: LEVel 0.16 /*Set the trigger level to 160mV*/
:TRIGger:EDGe: LEVel? /*The query returns 1.600e-1*/





4.7 TRIGger: PULSe
Command List
l TRIGger: PULSe:SOURce
l TRIGger: PULSe: POLarity
l TRIGger: PULSe:WHEN
l TRIGger: PULSe:WIDth
l TRIGger: PULSe: LEVel


4.7.1 TRIGger: PULSe:SOURce

Command
:TRIGger: PULSe:SOURce <source>
:TRIGger: PULSe:SOURce?
Description
Set or query the trigger source of the pulse width trigger.
Parameters
:<source> ::= < CHANnel1 | CHANnel2 | CHANnel3 | CHANnel4>
Return
The query returns CHANnel1, CHANnel2, CHANnel3, CHANnel4
Example
:TRIGger: PULSe:SOURce CHANnel1 /*set trigger source to CH1*/
:TRIGger: PULSe:SOURce? /*The query returns CHANnel1*/






58

Threads

1268

Posts

1348

Credits

金牌会员

Rank: 6Rank: 6

Credits
1348
30#
 Author| Post time 2021-10-9 08:16:42 | Show all posts
Edited by WisdomAugust at 2021-10-9 08:23

4.7.2 TRIGger: PULSe: POLarity
Command
:TRIGger: PULSe: POLarity <polarity>
:TRIGger: PULSe: POLarity?
Description
Set or query the trigger polarity of pulse width trigger
Parameters
:<polarity> ::= < POSItive | NEGAtive>
State
POSItive: Positive pulse trigger
NEGAtive: negative pulse trigger
Return
The query returns POSItive, NEGAtive
Example
:TRIGger: PULSe: POLarity POSItive /*Set positive pulse trigger*/
:TRIGger: PULSe: POLarity? /*The query returns POSItive*/

4.7.3 TRIGger: PULSe:WHEN
Command
:TRIGger: PULSe:WHEN <when>
:TRIGger: PULSe:WHEN?
Description
Set or query the trigger conditions for pulse width trigger.
Parameters
<when> ::= < EQUAl| NEQUal| GREAt| LESS>
State
?EQUAl = (equal to time value): Only when the positive or negative pulse width
of the input signal is equal to the set pulse width can it trigger. [Pulse width error is 5%] NEQUal! = (Not equal to time value): Only when the positive or negative pulse width of the input signal is not equal to the set pulse width, can the [pulse
width error be 5%] be triggered. GREAt > (greater than the time value): Only when the positive or negative pulse width of the input signal is greater than the set pulse width can it trigger [pulse width error is 5%].
?LESS <(less than time value): Only when the positive or negative pulse width of
the input signal is less than the set pulse width can the [pulse width error be
5%] be triggered.
Return
The query returns EQUAl, NEQUal, GRAt, LESS
Example
:TRIGger: PULSe:WHEN NEQUal /*Set trigger condition to NEQUal*/
:TRIGger: PULSe:WHEN? /*The query returns NEQUal*/





You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-8 06:50 GMT+8 , Processed in 0.273242 second(s), 20 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list