Electronic Engineer Discuss

View: 2498|Reply: 1
Print Prev. thread Next thread

DSO2C10 Signal Display Error

[Copy link]

1

Threads

2

Posts

2

Credits

新手上路

Rank: 1

Credits
2
Jump to specified page
1#
Post time 2023-7-22 01:39:16 | Show the author posts only Reply Awards |Ascending |Read mode
Edited by Z_E_V_S at 2023-7-22 01:33

I bought a Hantek DSO2С10 SCOPE (Number CN2302029034653 software 1.0.2.0.0 (221028.00) firmware 3205).
Detected a signal display problem.
If the signal frequency is 3 hz, or less, then the trigger does not work in auto mode.



If the trigger switches to normal mode, it works.



If you change the vertical resolution from 1 to 2 volts per division, the oscilloscope shows the correct signal for a short time.



But then it shows a value 2 times less



If you switch the trigger to auto mode, the oscilloscope shows the correct signal amplitude, but the trigger does not work.



If you now switch the trigger to normal mode, then the oscilloscope shows everything correctly.
There is also a problem with signal mapping when changing the depth of memory.
If you set the horizontal resolution of 50 ms per division and depth of the 400k memory, the oscilloscope displays the signal correctly.



If you switch the horizontal resolution by 100 ms per division, then the oscilloscope shows the wrong waveform



If you now press the single button, the oscilloscope shows the correct waveform



In the oscilloscope, it is not possible to reset the measurement statistics.
You need to reset everything, then reinstall the necessary measurements and turn on the statistics.
It will be better if you replace "Clear all" with "Clear statistics."

I really hope that in the next software change, these problems will be fixed.





This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
Reply

Use magic Report

1

Threads

2

Posts

2

Credits

新手上路

Rank: 1

Credits
2
2#
 Author| Post time 2023-7-30 21:08:51 | Show the author posts only
Hello.
Today I found new software on your website.
Installed this update.
This is the new version 1.0.3.0.0 (20230327.00).
Unfortunately, the bugs I described have not been fixed in the new software.
Error occurred when switching horizontal resolution from 50ms. at 100ms. with a memory depth of 400k or more, this is the biggest error.
Dear Hantek Engineers, you produce MEASURING equipment, and in this case your device is lying, and you do not want to correct the program error.
I ask you to answer, will these errors be corrected in the next update?

Maybe this is only in my oscilloscope such a mistake?
Someone, please check this error in your oscilloscope.


You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-11 22:47 GMT+8 , Processed in 0.157509 second(s), 22 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list